Have to agree.As you might have guessed, it's utter crap..but was fun to play with:
You never know until you try.
Have to agree.As you might have guessed, it's utter crap..but was fun to play with:
Good thought!And that's also probably why it's not linear (the resistor value is to high).
Oh wow... sure is! Sorry about that... .255 nF on the leads of the CT as measured by the meter without any primary load.And 2.55 F? unless there's a supercap somewhere I guess it's a typo ^^ also, it's the capacitance between what and what?
Thanks! The internal resistance numbers still don't really make a whole lot of sense to me....The secondary 145 ohm internal resistance is not really relevant to anything, except for estimating power dissipation and internal temperature rise.
I was thinking about that too, I see how it can be done if the saturation is high enough.Wide range measuring instruments will almost certainly use an amplifier between the CT and any analog to digital converter, and the gain might also be switched for autoranging.
The voltages were quite slow to steady, on the order of seconds. Not sure how much of that is the CT capacitance and how much is the meter, but guessing a bit of both. As you can see, at 100Ω the accuracies weren't the best.
I was thinking about that too, I see how it can be done if the saturation is high enough.
Guessing it probably makes sense to get a precision ADC than a CT with a higher saturation. Pity, a lot of microcontrollers have spare ADC pins, but it's doubtful they're accurate under a mV.
CTs have saturation voltages, I believe that's the magnetics, so the closer you get to it the more inaccurate the readings probably are. But the prior results are pretty interesting so that might be a datasheet error or just a coincidence.
CTs have capacitance. Again, it might be the meter; but, if you need something that measures very rapid changes (e.g., fingerprinting like the Sense energy monitor), you probably want less capacitance (e.g., fewer windings or some other trick).
I know Warpseed said the maximum burden resistor value was 100Ω based on the datasheet, but I don't think the CT cares as long as the voltage is well below the saturation.
The CT "quarantine" is 3 KV, but I've no idea what that is about. Possibly they mean the insulation's dielectric strength won't hold over 3KV?
Update: Like Warpspeed in his post above, I believe this to be a translation error.
The voltage isn't the source of the problem, it's the core which saturates if the magnetic field is too high.
What actually happens when the core saturates? From WIkipedia:
That description matches pretty well with the results in #62. Different core materials have different falloffs. From what I've read silicon steel is the preferred core material for CTs. But, there are two big takeaways for me from the curves. First is that a CT should be able to be operated pretty close to its saturation without sacrificing much accuracy. The second is that readings beyond it aren't total gibberish. It is still fairly linear past the saturation point although the flatter curve means a lot less sensitivity. |
Thanks! I scratched that thought out in my earlier conclusion, I may have been influenced by past calculations for capacitor charging times.Given it's seconds here I'd say it comes from the meter (did you have some averaging function enabled?). The CT should be good up to at least the kHz region giving it a less than 1 ms response time, far more than enough for typical applications.
No calibration, just raw data from which it could be calculated. The "err" column is (theoretical-actual)/theoretical.Did you do a point-slope linear calibration?
That's a really good idea!Or an op-amp.
I should probably start a thread on minimizing noise.At work we're playing with circuits that measure uA ripple in A circuits, and amplify it to where we can capture with DC coupled scope, having noise floor of maybe 0.1 mV. When I used a current probe on that scope, it had noise floor of 1 mA (corresponding to 1 mV into scope.)
Given the nature of the beast (measuring current through external circuits) I'm not sure how possible that is.Regarding noise, EMI, etc... I'd recommend to amplify as close to the CT as possible (an op-amp is enough) so the SNR is higher on the rest of the circuit.
I'm not sure what do you think is not possible?Given the nature of the beast (measuring current through external circuits) I'm not sure how possible that is.
Yep, shielded twisted pair is probably the best here without going crazy on complexity and costAlthough, shielded Wires might be a good requirement for a CT.
For example, on the Iotawatt the CT cables are 5'. They have to be long enough to string from whatever group of wires you want to take the reading back to the device that takes the voltage and reports it.I'm not sure what do you think is not possible?
What actually happens when the core saturates?
Thanks! I scratched that thought out in my earlier conclusion, I may have been influenced by past calculations for capacitor charging times.
What effects would capacitance have if any then? Other than as filters, capacitors in AC circuits are always a little mind-bending for me.
Hmm, is that what the datasheet is trying to say by phase difference, which they specify as <5 (20A, 0Ω) Branch? I suspect another translation error. But possibly they mean the delay is <5 degrees of the phase? Which would be 0.23 mS?
No calibration, just raw data from which it could be calculated. The "err" column is (theoretical-actual)/theoretical.
The thing that immediately hits me when using an op-amp or precision ADC is does it really have any value to do so? In my test case, the length of the wires, using a wago, proximity of the neutral, proximity to a PC... any of that can start to alter voltage measurements in the sub mV range.
But anyway, since the CT by its nature would most likely be distant from the measuring apparatus, probably just as important is the circuitry to eliminate unwanted AC frequencies picked up. Not sure how you'd do that as devices like the IotaWatt would be deployed inside the breaker box where all the wires are radiating at the same frequency.
The accuracy of the CTs I tested was listed as 0.1, but no units so it's meaningless.
Regarding noise, EMI, etc... I'd recommend to amplify as close to the CT as possible (an op-amp is enough) so the SNR is higher on the rest of the circuit.
Regarding accuracy the general rule is that a percentage means it's a fraction of the measured value, not a fraction of the meter maximum. Other units (V, A, digits, ...) usually means it's a constant value regardless of the measured value.
Driven current should be fairly low impedance, more immune. Following burden resistor, converted to small voltage more sensitive.
Run wires from current transformer, put burden resistor fairly close to amp or ADC?
My first thought was that "0.1" meant "10%"
Many sensors also have an offset, but don't think there would be for CT.